READINGS TOO HIGH OR TOO LOW
Article Details

Last Updated
21st of March, 2012

User Opinions (7 votes)
100% thumbs up 0% thumbs down

How would you rate this answer?
Helpful
Not helpful

Calibration may be off.  The calibration procedure is published in the manual and copied below.


1. Balance the meter in an upright position on a hard surface. 

2. Press and hold the START button.

3. With the START button depressed, press and release the ON button.  The LCD will count down from 5 to 1.  

4. The meter should now read between 5 and 10 P.S.I. (35 - 70 Kpa).  This reflects the weight of the meter itself.


Note: During the calibration procedure, care must be taken to ensure that the only force on the load cell is the weight of the meter itself.

 